why water, methanol, ethanol ets solvents are used in UV
SPECTRIPHOTOMETER for blank reading ?
Answers were Sorted based on User's Feedback
Answer / chemdeep
usually the same solvent is used in uv for blank readings
that is used in the sample preparation.
Is This Answer Correct ? | 28 Yes | 6 No |
Answer / rama
If your medium is given solvents, then that solvents can be
used as blank other wise that can't be.
Is This Answer Correct ? | 17 Yes | 4 No |
Answer / nitesh pal
all that solvent not absorb uv light i.e not show absorbance in uv region
Is This Answer Correct ? | 3 Yes | 2 No |
Answer / nabil
Methanol for example absorbs at wavelengths below nm, and
water below 190 nm if we use methanol-water mix as a blank
we would have to use a wavelength grater than 205 nm to
avoid false reading.
Is This Answer Correct ? | 5 Yes | 8 No |
Answer / radha
they do not give any reaction with sample did n't give peaks
Is This Answer Correct ? | 7 Yes | 15 No |
why hexadecane is used in calibration of FID IN GC THERE IS NO REMAINING SOLVENTS
difference between pH & pOH ?
What is the difference between Spectroscopy and spectrophotometry,What are the applications?
4 Answers Caplin Point, Cipla,
Why are use silicon oil mr?
Why dissolution test apparatus calibration with salycilic acid tablets was stopped by FDA&USP from the year of 2009?
How to decide assay range for non pharmacopeial API analysis by HPLC? Could you give me any reference for same e.g Guidelines or Paper publications?
what is the difference between c-8 and c-18 column
What is Impuriteis?
in gas chromatography what is the difference between gas flow rate and average linear velocity ?
in calibration of disintegration test apparatus why cycle limit is 28-32?
in dissolution in one tablets two molecule one is losartan & second is HCTZ losartam is complies in s1 stage & hctz is complies in s2 stage then how to report result losartan s1 stage result or losartan s2 stage result
why urasil use in hplc column perfomance?