What is the procedure for UV-spectrophotometer calibration?

Answers were Sorted based on User's Feedback



What is the procedure for UV-spectrophotometer calibration?..

Answer / manikatna

UV-VIS spectrophotometer calibration is many parameters
which is
1)control of wavelength
2)control of absorbance
3)wavelength accuracy
4)straylight accuracy
5)resolution
6)resolution power

Is This Answer Correct ?    172 Yes 15 No

What is the procedure for UV-spectrophotometer calibration?..

Answer / rakesh

1.0 CONTROL OF ABSORBANCES
PROCEDURE: REPARATION OF POTASSIUM DICROMATE SOLUTION: Dry
K2Cr2O7 at 130oC for 2 hours. Weigh accurately about the
quantity of sample equal to 60mg into 1000ml volumetric
flask. Add 0.005m Sulphuric acid to dissolve and dilute
with the same solvent.
Find out the absorbances of (A 1% 1cm) above said solution
at wavelengths of a 235nm, 257nm, 313nm and 350nm.
Calculate the (1% 1cm) value using the following expression.
Absorbance X 1 X 1000 / Weight of K2Cr2O7 in g X 100
Check the value for each wavelength with the tolerances
given four wavelengths.


PREPARATION OF REAGENT:
(i) Dilute 3ml of concentrated sulphuric acid and make up
100ml with distilled water.
(ii) Dilute 20ml to 2000ml with distilled water.
2. LIMIT OF STRAYLIGHT:
The absorbance of a 1.2-% w/v solution of potassium
chloride at a path-length of 1cm should be greater than 2.0
at about 200nm when compared with water as reference liquid.
1.2% POTASSIUM CHLORIDE SOLUTION:
Take 1.2 g of Potassium chloride previously dried at 105oC
for two hours into the 100ml volumetric flask. Add water to
dissolve and make up to volume with water.
3. RESOLUTION POWER:
Record the spectrum of a 0.02% v/v solution of Toluene in
Hexane UV. The ratio of the absorbance at the maximum at
about 269nm to that at the minimum at about 266nm is not
less than 1.5 unless other wise specified in the monograph.
0.02% v/v TOLUENE:
Take 0.02ml of Toluene into the 100ml volumetric flask. Add
Hexane to volume.

Is This Answer Correct ?    97 Yes 8 No

What is the procedure for UV-spectrophotometer calibration?..

Answer / upk

UV-VIS spectrophotometer calibration is many parameters
which is
1)control of wavelength
2)control of absorbance
3)wavelength accuracy
4)straylight accuracy
5)resolution
6)resolution power &
7) Io Line Flatnes test

Is This Answer Correct ?    55 Yes 22 No

What is the procedure for UV-spectrophotometer calibration?..

Answer / nitesh pal

in above answer also added 430 nm in control of wavelength.(acc to ipm dilution is about 60 mg to 1ooml (same solvent)

Is This Answer Correct ?    21 Yes 8 No

What is the procedure for UV-spectrophotometer calibration?..

Answer / ramkrishna jagtap

control of wavelength

Is This Answer Correct ?    7 Yes 0 No

What is the procedure for UV-spectrophotometer calibration?..

Answer / devender justa

Internal Calibration ,external Calibration, Baseline
Flatness

Is This Answer Correct ?    8 Yes 4 No

Post New Answer

More Analytical Chemistry Interview Questions

What is (Q) point dissolution?

6 Answers   Cipla, Macleods,


we can use expired sample for validation and analyst qualification?

0 Answers  


in loss on drying Why follows on constant weight not more than 0.5 mg, any region

1 Answers   Biocon,


Column selection in HPLC method development?

1 Answers  


What is the difference between Normal phase method & Reverse method in HPLC ? Define application of both method.

4 Answers  






what is the maximum flow rate in prep hplc

1 Answers   Meril Life Science,


in DMF having extra impurities and in api COA also having extra imp than USP or BP product then how require to proceed?

0 Answers  


why kbr is used for pellet preparation in ir spectroscopy, give the reasons? why kbr shows zero dipole moment?

6 Answers  


[3/30, 13:29] Manoj P Venkatpurwar: How hplc column selection according to structure? How mobile phase buffer selection on molecule structure?

0 Answers  


In HPLC when we use Isocratic method, Gradient method in hplc?

8 Answers   Pharmaceutical,


some compound havving uv region but not give much more response during hplc analysis

2 Answers  


what parameters we will consider while developing a HPLC method and how we confirm our HPLC method is valid?

0 Answers   DRL, Laurus,


Categories
  • Organic Chemistry Interview Questions Organic Chemistry (342)
  • Inorganic Chemistry Interview Questions Inorganic Chemistry (123)
  • Analytical Chemistry Interview Questions Analytical Chemistry (1387)
  • Physical Chemistry Interview Questions Physical Chemistry (62)
  • General Chemistry Interview Questions General Chemistry (478)
  • Chemistry AllOther Interview Questions Chemistry AllOther (190)