why linerity not use in GC but use in HPLC ?
Answers were Sorted based on User's Feedback
Answer / mr.bujji reddy kanchi
Linearity can be used in GC also.
Lineariy must be evaluated in case of analytical method
validation and calibration.
Is This Answer Correct ? | 18 Yes | 3 No |
Answer / d.sambireddy
Linearity also using in gc
In Any Analytical method validation and calibration of
GC. Linearity is one of the Importent parameter
Is This Answer Correct ? | 4 Yes | 0 No |
Answer / jangam shyam
linearity test used in hplc and gc.this test is performed to
detect the performance of detector.if we increase the conc
of the sample vise versa the response is increased.the graph
shold be linear.
Is This Answer Correct ? | 3 Yes | 0 No |
Answer / sandip bhasme
If you are taking % area of a componant; there is no need to
take linearity in GC, But if you have to compare standard
against sample, you need to check linearity for GC also. eg.
if you are analyzing a gas sample on FID OR TCD detector,
your detector response must be linear.
Is This Answer Correct ? | 2 Yes | 0 No |
Answer / madhu
linearity is one of the main importent calibration parameter in GC& hplc
Is This Answer Correct ? | 0 Yes | 0 No |
Answer / nani
Linearity should be evaluated GC.
1.Detector responce linearity
it shows your dector linea or not
Is This Answer Correct ? | 0 Yes | 1 No |
what r the reasons to get peak telling?
what is principal of karl ficher instrument!
difference between HPLC & LC-MS
what is deference b/w the working standard and reference standard
what is stability study ? why it is needed and how itis done
in calibration of disintegration test apparatus why cycle limit is 28-32?
How can v calibrate ftir ? Plz explain me
identification is for unknown? qualification for known? reporting for LOQ?
IN DISSOLUTION TEST S1AND S2 AND S3 CONDITIONS ALL ARE FAILED THEN WHAT IS THE DESSICION U CAN TAKE
How you will select column for HPLC development depends on nature of compound and how it will work explain in brief?
why IR graph or spectra comes down ? Any reason
When you get Moisture content by KF higher than by LOD, what does this indicate?