how FTIR calibration Doing?
Answers were Sorted based on User's Feedback
Answer / murugesan
with polystyrene film
wave number check and resolution as per BP
| Is This Answer Correct ? | 19 Yes | 1 No |
Answer / ramkrishna thakare
1)with polystyrene film
2)wave number check
3 ) resolution.
4) energy efficiency.
Ramkrishna thakare --------Sandoz
| Is This Answer Correct ? | 11 Yes | 2 No |
Answer / ramu
Doing the calibration of FT-IR using Polystyrene film.You
have to do the three things in the calibration procrdure.
1)Verification of Wave number scale
2)Control of resolution between two types you have to check
a)Transmission minimum at 1583
Transmission maximum at 1589
b)Transmission minimum at 2849.5
Transmission maximum at 2870
| Is This Answer Correct ? | 9 Yes | 1 No |