acceptance criteria for ph meter, hplc
detector,pump,injector,oven when we calibrate.

Answers were Sorted based on User's Feedback



acceptance criteria for ph meter, hplc detector,pump,injector,oven when we calibrate...

Answer / guest

Acceptance Criteria
---------------------

pH Meter:
-----------

Perform 2 point or 3 point calibration as per the
use/requirement.

The slope should be 95% to 105%

Where as pH measured value for known pH buffer should be
4.01 ± 0.05
7.00 ± 0.05
9.21 ± 0.05
10.00 ± 0.05
11.00 ± 0.05

HPLC Detector:
--------------

Prepare differenct concentration of caffeine solution say
10 ppm, 25 ppm, 50 ppm, 75 ppm, 100 ppm, and 125 ppm.

Inject single injection of each solution and calculate the
correlation coefficient = sholud be 0.999

HPLC Pump:
----------

a) Flow rate accuracy
b) GPV

a) Flow rate accuracy:

Calculate flow rate accuracy using water at difference flow
rate at 0.5 mL/min , 1.0 mL/min, and 3.0 mL/min at duration
of 5 minutes and calculate the volume collected with in 5
minutes duration using stop watch collected in pre-weigh
calss A volumetric flask.Apply density of water at
specified temperature.

Acceptance Criteria: For all the flow rate should be ± 5%

b) Gradient Proportianating Value (GPV)

GPV test will be perform using 5.6 mg of Propyl paraben ----
-} 1000 mL of methanol for channel A and B if we are
calibrating the Quadranaery pump and C and D channels in
methanol run the gradient composition at the specified time
to obtain the response when compared with higest response
of methanol vs propyl paraben solution should be 10 ± 0.5
i.e 9.5 - 10.5.

We can even calibrate using 0.5% Acetone in water by
evaluating the response obtained throughout the run time.

Injector:
---------

a) Injector Linearity
b) Injector precision
c) Volume accuracy

a) Injector Linearity

Inject singe injection of 25ppm caffeine solution of
injection volume 5µ, 10µ, 20µ, 50µ and 100µ. if the
injector loop is 100 µl loop where as if the injector loop
is 200 µl then go upto 200 µl for better accuracy.

Acceptance criteria: Correlation coefficient: 0.999

b) Injector Precision:

Inject 5 times of 25 ppm caffeine solution and calculate
the RSD for area.

Accetance criteria: < 1.0%

c) Volume Accuracy:

Check the injector volume accuracy by intecting the 5µl
volume for 10 times in a pre-weigh vial containing water.

Check the injector volume accuracy by intecting the 10µl
volume for 10 times in a pre-weigh vial containing water.

Acceptance criteria: 4.95 - 5.05 µl
19.5 - 20.5 µl

Column Oven:
------------

Calibrate the column oven with calibrated
thermometer/digital at 30° C and 70°C

Acceptance Criteria: 30 ± 2 °C
70 ± 2 °C

Is This Answer Correct ?    19 Yes 1 No

acceptance criteria for ph meter, hplc detector,pump,injector,oven when we calibrate...

Answer / saptale maruti

Flow rate ± 0.2% of set value
Flow rate consistancy RSD NMT1.0%
Compositional Accuaracy : Difference in Absorbance is 0.01AU
Difference in time NMT 20SEc
Delay Volume NMT 4.5 (For 4 ml Delay Volume System)
Injection Volume Accuracy is : Average volume for 18
injection is 19.6 to 20.4µl
Linearity 0.999
Wavelenght Accuracy Maxima 273± 2 nm
Minima 205 and 245± 2 nm
Oven Temperature 2 degree from set temp.

For Ph ± 0.02

Is This Answer Correct ?    7 Yes 2 No

acceptance criteria for ph meter, hplc detector,pump,injector,oven when we calibrate...

Answer / basha

For pH + OR - 0.05
HPLC-DETECTOR + OR - 1nm
Acceptance Criteria: For all the flow rate should be ± 0.02
Acceptance criteria: Correlation coefficient: 0.98
Accetance criteria for volume:1 %
Acceptance Criteria: ± 1 °C

Is This Answer Correct ?    6 Yes 7 No

Post New Answer

More Analytical Chemistry Interview Questions

What is the Formula for coreletion coefficient in plhplc calibration

0 Answers  


why we are using sodium,potassium phosphates for mobile phase preparation?

1 Answers  


Why sodium lamp is used as source of Polarimeter?

12 Answers   College School Exams Tests, Hetero, NCL, Zydus Cadila,


Which water can use Hplc analysis

9 Answers   BioGenomics, Dr Reddys, Reddy Labs,


As per ICH related substances stability trend limit from initial to shelf life

0 Answers   Medley Pharmaceuticals, Micro Labs,






What is the Baseline in HPLC.

2 Answers   Teva API,


What is difference between NDA & ANDA ?

1 Answers   Alembic,


%lod always grater than %water content for same product??

1 Answers  


How can we calculate offset value pH measurment

0 Answers   Dr Reddys,


use of stry light

1 Answers  


How we develop a method by HPLC for unknown compound,if we got three peaks with same response then how we can find out which one is our desired compound's peak without taking help of LC MS.

4 Answers   Apotex, Kemwell,


If change in specification which parameter require to do for validation? if change in chromtographic condition then which parameter? if api change then which parameter? if change composition then which parameter? if old method not work out then whicj parameter? if additional one impurity added then which parameter of validation require to do on above each conditions? elaborate separately

0 Answers  


Categories
  • Organic Chemistry Interview Questions Organic Chemistry (342)
  • Inorganic Chemistry Interview Questions Inorganic Chemistry (123)
  • Analytical Chemistry Interview Questions Analytical Chemistry (1387)
  • Physical Chemistry Interview Questions Physical Chemistry (62)
  • General Chemistry Interview Questions General Chemistry (478)
  • Chemistry AllOther Interview Questions Chemistry AllOther (190)